Bench "Standard" Check Up
The hp 3456A is my bench voltage/resistance bench "standard" so it's time to check it's accuracy against some known references. First up is DC volts. Primary reference is an AD584-M that was checked at the factory with an Agilent 34401A at 21 C ambient. Both this reference and the 3456A were powered up for 6 hours at 22 C ambient. Results as follows:
AD584-M.............3456A
2.50066V.............2.50061V
4.99794V.............4.99791V
7.50011V.............7.50010V
9.99691V.............9.99690V
Next up is the secondary AD584-M reference. That was checked at the factory with an Agilent 34401A at 25 C ambient. The 3456A was powered up for 9 hours and the AD584-M powered up for 3 hours. Results as follows:
AD584-M.............3456A
2.49937V.............2.49933V
5.00030V.............5.00026V
7.50042V.............7.50039V
10.00096V...........10.00094V
Based upon these results the 3456A has experienced basically zero drift on DCV since the last time it was checked.
I don't have a high accuracy ACV reference other than a 1.00VAC which I wouldn't trust beyond 2 decimal places. When applied to the 3456A it reads 1.018286V. Good enough for my needs.
Next up is resistance. Leeds and Northrup 4747. This resistance box is showing some wear and tear so I had to do a lot of cleaning and futzing with it. The 2 binding posts are corroded and initially had a contact resistance of about 0.0450 ohms. With cleaning I got it down to 0.0148 ohms. But future I'll be replacing those binding posts with high quality pieces. Also on the 100 and 1000 ohm ranges some additional contact resistance is present. The switch decks show considerable wear and corrosion that deoxit won't touch. I'm considering using some metal polish. Future project. The 3456A had been powered up for nearly 12 hours. So the results are as follows:
4747...............3456A
1.0.................1.0159
10.0...............10.0191
100.0.............100.0629
1000.0............1000.317
The 1 and 10 ohm are OK while the 100 and 1000 are a little suspect but I think it's somewhat safe to conclude that the 3456A is OK.
Tomorrow will check my 10MHz "standard".
AD584-M.............3456A
2.50066V.............2.50061V
4.99794V.............4.99791V
7.50011V.............7.50010V
9.99691V.............9.99690V
Next up is the secondary AD584-M reference. That was checked at the factory with an Agilent 34401A at 25 C ambient. The 3456A was powered up for 9 hours and the AD584-M powered up for 3 hours. Results as follows:
AD584-M.............3456A
2.49937V.............2.49933V
5.00030V.............5.00026V
7.50042V.............7.50039V
10.00096V...........10.00094V
Based upon these results the 3456A has experienced basically zero drift on DCV since the last time it was checked.
I don't have a high accuracy ACV reference other than a 1.00VAC which I wouldn't trust beyond 2 decimal places. When applied to the 3456A it reads 1.018286V. Good enough for my needs.
Next up is resistance. Leeds and Northrup 4747. This resistance box is showing some wear and tear so I had to do a lot of cleaning and futzing with it. The 2 binding posts are corroded and initially had a contact resistance of about 0.0450 ohms. With cleaning I got it down to 0.0148 ohms. But future I'll be replacing those binding posts with high quality pieces. Also on the 100 and 1000 ohm ranges some additional contact resistance is present. The switch decks show considerable wear and corrosion that deoxit won't touch. I'm considering using some metal polish. Future project. The 3456A had been powered up for nearly 12 hours. So the results are as follows:
4747...............3456A
1.0.................1.0159
10.0...............10.0191
100.0.............100.0629
1000.0............1000.317
The 1 and 10 ohm are OK while the 100 and 1000 are a little suspect but I think it's somewhat safe to conclude that the 3456A is OK.
Tomorrow will check my 10MHz "standard".
Last edited by MED6753 on Wed Dec 07, 2022 2:48 pm, edited 1 time in total.
An old gray beard with an attitude. I don't bite.....sometimes
Tags:
- Specmaster
- Posts: 1143
- Joined: Sat Oct 22, 2022 8:13 pm
- Location: Chelmsford, UK
Re: Bench "Standard" Check Up
I think you need to double-check your calculations, I would say that the most accurate resistances you have are the 1000 and 100 as both are giving you readings of 99.96831% and 99.9371395% respectively while the 10 and 1 are providing you with 99.8093641% and 98.4348853% respectively.MED6753 wrote: ↑Tue Dec 06, 2022 5:03 pm Next up is resistance. Leeds and Northrup 4747. This resistance box is showing some wear and tear so I had to do a lot of cleaning and futzing with it. The 2 binding posts are corroded and initially had a contact resistance of about 0.0450 ohms. With cleaning I got it down to 0.0148 ohms. But future I'll be replacing those binding posts with high quality pieces. Also on the 100 and 1000 ohm ranges some additional contact resistance is present. The switch decks show considerable wear and corrosion that deoxit won't touch. I'm considering using some metal polish. Future project. The 3456A had been powered up for nearly 12 hours. So the results are as follows:
4747...............3456A
1.0.................1.0159
10.0...............10.0191
100.0.............100.0629
1000.0............1000.317
The 1 and 10 ohm are OK while the 100 and 1000 are a little suspect but I think it's somewhat safe to conclude that the 3456A is OK.
Tomorrow will check my 10MHz "standard".
Who let Murphy in?
Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Advance-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi-Heathkit-Duratool
Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Advance-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi-Heathkit-Duratool
Re: Bench "Standard" Check Up
You're failing to consider the binding post contact resistance of approx 0.0148 ohms which accounts for the apparent greater error on the 1 ohm and 10 ohm. If you subtract the binding post contact resistance you get 1.0011 ohm and 10.0043 ohms.Specmaster wrote: ↑Tue Dec 06, 2022 6:29 pm
I think you need to double-check your calculations, I would say that the most accurate resistances you have are the 1000 and 100 as both are giving you readings of 99.96831% and 99.9371395% respectively while the 10 and 1 are providing you with 99.8093641% and 98.4348853% respectively.
An old gray beard with an attitude. I don't bite.....sometimes
- Specmaster
- Posts: 1143
- Joined: Sat Oct 22, 2022 8:13 pm
- Location: Chelmsford, UK
Re: Bench "Standard" Check Up
Oops That would make a difference, either way the results are still pretty good.MED6753 wrote: ↑Tue Dec 06, 2022 9:33 pmYou're failing to consider the binding post contact resistance of approx 0.0148 ohms which accounts for the apparent greater error on the 1 ohm and 10 ohm. If you subtract the binding post contact resistance you get 1.0011 ohm and 10.0043 ohms.Specmaster wrote: ↑Tue Dec 06, 2022 6:29 pm
I think you need to double-check your calculations, I would say that the most accurate resistances you have are the 1000 and 100 as both are giving you readings of 99.96831% and 99.9371395% respectively while the 10 and 1 are providing you with 99.8093641% and 98.4348853% respectively.
Who let Murphy in?
Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Advance-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi-Heathkit-Duratool
Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Advance-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi-Heathkit-Duratool
Re: Bench "Standard" Check Up
Continuing with "Bench standards". 10MHz reference. Mine is a stand alone unit built up from a reference given to me by nixiefreqq back in 2019. I enclosed it in it's own case with power supply. It's only powered up when I need it so obviously it has to be on for several hours for best accuracy.
After several hours of up time on both the reference and monitoring frequency counters this is the result:
After running overnight the Star Trek Shuttle Fluke 7260A was showing +2 to +3Hz high. So time to break out the big guns and see what it says.
So after about 6 hours up time on the 5248L every other sample was either nominal or +1Hz.
Conclusion: This 10MHz reference is adequate for my needs and worst case is perhaps +3Hz high although the 7260A might be at fault here and need an adjustment. But I'm not going to sweat it and leave it as is.
After several hours of up time on both the reference and monitoring frequency counters this is the result:
After running overnight the Star Trek Shuttle Fluke 7260A was showing +2 to +3Hz high. So time to break out the big guns and see what it says.
So after about 6 hours up time on the 5248L every other sample was either nominal or +1Hz.
Conclusion: This 10MHz reference is adequate for my needs and worst case is perhaps +3Hz high although the 7260A might be at fault here and need an adjustment. But I'm not going to sweat it and leave it as is.
An old gray beard with an attitude. I don't bite.....sometimes
- nixiefreqq
- Posts: 159
- Joined: Fri Oct 28, 2022 12:41 pm
- Location: commonwealth of pennsylvania
Re: Bench "Standard" Check Up
for those youngsters new to direct counters (not you med):
"The value of the LSD Displayed often is the same as the quantization
error which represents the smallest non-zero change which may be
observed on the display. Because of this, resolution and accuracy
statements often specify quantization error as ± LSD Displayed."
"Although we say ±1 count, we do not mean that a particular measurement
can vary by both + and – one count. The measurement can vary
by one count at most. The reason that you have to say ±1 count is that
from a single measurement, you don’t know a priori which way the next
measurement will jump or if it will jump at all. So the specification has
to include both possibilities."
consider the 10000001 displayed to be a feature, not a fault......i still love direct count nixie counters and view reciprocal counters as cheating.
the 5248 is a thing of beauty until the tunnel diodes in the trigger go tits up. hmmmmm. i may have some trigger boards in my massive stash of 5245/48 parts.
"The value of the LSD Displayed often is the same as the quantization
error which represents the smallest non-zero change which may be
observed on the display. Because of this, resolution and accuracy
statements often specify quantization error as ± LSD Displayed."
"Although we say ±1 count, we do not mean that a particular measurement
can vary by both + and – one count. The measurement can vary
by one count at most. The reason that you have to say ±1 count is that
from a single measurement, you don’t know a priori which way the next
measurement will jump or if it will jump at all. So the specification has
to include both possibilities."
consider the 10000001 displayed to be a feature, not a fault......i still love direct count nixie counters and view reciprocal counters as cheating.
the 5248 is a thing of beauty until the tunnel diodes in the trigger go tits up. hmmmmm. i may have some trigger boards in my massive stash of 5245/48 parts.
free range primate since 2011
Re: Bench "Standard" Check Up
Oh wonderful. So hp gear uses unobtainium tunnel diodes like older Tek gear? Didn't know that. Knock on wood (taps side of head) I haven't (yet) had one go bad.nixiefreqq wrote: ↑Thu Dec 08, 2022 12:50 pm
the 5248 is a thing of beauty until the tunnel diodes in the trigger go tits up. hmmmmm. i may have some trigger boards in my massive stash of 5245/48 parts.
An old gray beard with an attitude. I don't bite.....sometimes
Re: Bench "Standard" Check Up
Having looked into it slightly, I don't get the impression that very many tunnel diodes are exactly made of unobtainium. There's some NOS as well as pulls. There's new production from AmericanMicrosemiconductor and others. The price? Well don't look if you have any history of a heart condition.
Then there are the Russkie ones still available on ebay from Ukraine and places such as Bulgaria. They seem reasonably priced apart from the postage and import duty. I've no idea what the Western equivalents are. I suppose someone has looked into it and there's a web page or two on them. I'm slightly tempted to get some to make a tunnel diode pulser.
Then there are the Russkie ones still available on ebay from Ukraine and places such as Bulgaria. They seem reasonably priced apart from the postage and import duty. I've no idea what the Western equivalents are. I suppose someone has looked into it and there's a web page or two on them. I'm slightly tempted to get some to make a tunnel diode pulser.
- nixiefreqq
- Posts: 159
- Joined: Fri Oct 28, 2022 12:41 pm
- Location: commonwealth of pennsylvania
Re: Bench "Standard" Check Up
maybe it is a possibility.......maybe not.MED6753 wrote: ↑Thu Dec 08, 2022 3:15 pmOh wonderful. So hp gear uses unobtainium tunnel diodes like older Tek gear? Didn't know that. Knock on wood (taps side of head) I haven't (yet) had one go bad.nixiefreqq wrote: ↑Thu Dec 08, 2022 12:50 pm
the 5248 is a thing of beauty until the tunnel diodes in the trigger go tits up. hmmmmm. i may have some trigger boards in my massive stash of 5245/48 parts.
thought about it some more and i seem to remember that only the early 5248's used tunnel diodes on the A17 and A21 boards.
but both of mine have dip ecl chips on those boards and the parts list does not show any tunnel diodes. but some parts lists definitely show tunnel diodes.
no idea what is in your 5248.
You do not have the required permissions to view the files attached to this post.
free range primate since 2011
Re: Bench "Standard" Check Up
I have no idea either and I really don't want to go looking for trouble. It's working, leave it alone. But I do wish there was an easy solution for that god awful noisy fan.nixiefreqq wrote: ↑Thu Dec 08, 2022 9:16 pm
but both of mine have dip ecl chips on those boards and the parts list does not show any tunnel diodes. but some parts lists definitely show tunnel diodes.
no idea what is in your 5248.
An old gray beard with an attitude. I don't bite.....sometimes
- nixiefreqq
- Posts: 159
- Joined: Fri Oct 28, 2022 12:41 pm
- Location: commonwealth of pennsylvania
Re: Bench "Standard" Check Up
no doubt you already tried all the obvious fixes for the fan. (a little 3 in one oil, or just clean the dust off). in my experience sometimes the blades have gotten bent out of alignment and a little careful tweaking gets them back in balance.MED6753 wrote: ↑Fri Dec 09, 2022 11:26 amI have no idea either and I really don't want to go looking for trouble. It's working, leave it alone. But I do wish there was an easy solution for that god awful noisy fan.nixiefreqq wrote: ↑Thu Dec 08, 2022 9:16 pm
but both of mine have dip ecl chips on those boards and the parts list does not show any tunnel diodes. but some parts lists definitely show tunnel diodes.
no idea what is in your 5248.
cub might have a fix for the noisy fan problem. (but he will probably just suggest that you get a cat).
free range primate since 2011
Re: Bench "Standard" Check Up
Naaahhh... he just needs a redhead to take his mind off those parts per million.
mnem
mnem
You do not have the required permissions to view the files attached to this post.