190VDC Reference Revisited.

The place to be when you have TEA. Discuss all kinds of test equipment.

Important: Use tags for the type of equipment your topic is about.
Forum rules
Use tags for the type of equipment your topic is about. Include the "repairs" tag, too, when appropriate. If a new tag is needed, request one in the TEAdministration forum.
Post Reply
User avatar
MED6753
Posts: 892
Joined: Sun Oct 23, 2022 6:02 pm
Location: Middletown, NY USA

190VDC Reference Revisited.

Post by MED6753 »

Finally got some lab time after nearly 2 months of doctor's appointments and other assorted distractions. Yes, Blondie is doing well. Revisiting a 190VDC reference that I built up several years ago. It was never very stable and in fact quit working all together. Discovered that the Inverter card would spike to over 270VDC at initial turn on which blew out the LR-8 regulator. This was solved by adding a 100K/2 watt resistor across the Inverter output. Replaced the LR-8 regulator and now have a much more stable output.

Schematic
Image

Internal view.
Image

External view.
Image

Results in low res mode.
Image

Results in high res mode.
Image

Cross check on a 8800A
Image

The reference was allowed to stabilize for about 45 minutes. I would consider it accurate enough to check/calibrate 3.5 and 4.5 digit DMM's but not something like the 3456A. So now I have a complete set of 190mV, 1.9V, 19V, and 190V references.