Beware PSU current limits...

The place to be when you have TEA. Discuss all kinds of test equipment.

Important: Use tags for the type of equipment your topic is about.
Forum rules
Use tags for the type of equipment your topic is about. Include the "repairs" tag, too, when appropriate. If a new tag is needed, request one in the TEAdministration forum.
User avatar
EC8010
Posts: 628
Joined: Mon Sep 08, 2025 12:41 am

Beware PSU current limits...

Post by EC8010 »

I have become used to the idea that if I want a constant current from a bench supply, I just dial the current limit to the required current then twiddle voltage to be adequate. This works on my Agilent E3648A and E3649A, and on most TTi supplies. It's a very nice way of working. A Keithley 2400 is different. Actually, everything about a Keithley 2400 is different, inconsistent, or just plain awkward. But if you can get over that, it's very capable. With a 2400, you have to explicitly tell it to be a voltage or current source, then you set compliance. And that turns out to be better...

The picture shows three stacked tag boards carrying a total of ninety nine Zeners. I have traditionally tested that I've wired such things correctly by setting the 2400 to 3mA and compliance to 210V. I can then check that I get the expected voltages at various points. But a 2400 only goes to 210V and I needed 300V, so I stacked a pair of TTi PLH250 and set the current limit to 3mA. This has worked fine before. But today, I didn't gently wind the voltage up, I dabbed it whilst watching the voltmeter. It turns out that the output capacitor (charged to umpty-ump volts) is capable of converting an entire series string of thirty eight Zeners into a short-circuit. That was an entire tag board that had to be thrown away and remade. Naturally, it was the bottom board (most difficult to get at) that was totally immolated, plus five more on the middle board. I was not pleased.

So, be warned. If you want to use a bench supply in constant current mode to test Zeners, set the current limit and gently wind the voltage up from zero. Do not dab it or switch the output on.

The remade (and very carefully checked) stack has gone into a die-cast aluminium box, been filled with white oil, had silicone sealant intended for engines applied to mating surfaces, and lid screwed down. I'll give it a day or so for the sealant to go off before disturbance and measurements.

Oh, and the feed through connectors are at Stage 2 of the dreaded "Won't solder" disease. When scraped back to bare brass, they just about soldered, but they didn't wet properly. Naturally, these are the feed throughs that I have plenty of. I have seen valve socket, solder tags, and connectors that have caught this disease. It's always tinned brass that is afflicted. Has anyone else seen this?
You do not have the required permissions to view the files attached to this post.

Tags:
tggzzz
Posts: 2841
Joined: Sat Oct 22, 2022 8:17 pm

Re: Beware PSU current limits...

Post by tggzzz »

Oh, yes indeed :(

https://entertaininghacks.wordpress.com ... behaviour/ gives examples of that. Some PSUs behave well, but it shows examples where a current limited PSU delivers 170 times the set power, current limits are ignored for 500ms, and don't reach the correct value for 1500ms.

3000µF output capacitance doesn't help!
User avatar
Specmaster
Posts: 1233
Joined: Sat Oct 22, 2022 8:13 pm
Location: Chelmsford, UK

Re: Beware PSU current limits...

Post by Specmaster »

Yes, I often see that, especially after a battery has let go, even scrapping the tag, track, etc., all the way back to bare copper. I have just ordered some special flux from AliExpress to see if that can help in that dept. It is supposed to even work on stainless steel, so here hoping, fingers crossed.
Who let Murphy in?

Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Advance-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi-Heathkit-Duratool
User avatar
EC8010
Posts: 628
Joined: Mon Sep 08, 2025 12:41 am

Re: Beware PSU current limits...

Post by EC8010 »

Please do report back if the flux works. Beware that flux is deliberately corrosive, so you're going to need to thoroughly wash it away...
User avatar
Specmaster
Posts: 1233
Joined: Sat Oct 22, 2022 8:13 pm
Location: Chelmsford, UK

Re: Beware PSU current limits...

Post by Specmaster »

Yes, I will do that. I'm also waiting on some magical solution that also removed all battery corrosion and is supposed to return contacts, etc., to their former glory. Sounds too good to be true, but we shall see. I have it on good authority that these items are now at a local airport and going through customs. Fingers crossed, it won't be long before we can find out if they work.
Who let Murphy in?

Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Advance-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi-Heathkit-Duratool
User avatar
EC8010
Posts: 628
Joined: Mon Sep 08, 2025 12:41 am

Re: Beware PSU current limits...

Post by EC8010 »

Specmaster wrote: ↑Wed Nov 18, 2026 2:46 pm I'm also waiting on some magical solution that also removed all battery corrosion and is supposed to return contacts, etc., to their former glory.
Fred Wedlock had a song about a cleaner that was effective on anything from shot silk to rusty armour plate, but the company making it couldn't think of a name, so they put it out as a staff competition. One of the lads suggested that it be called "Best Universal Grit, Grime, and Effluent Remover," and added that a suitable slogan would be, "If Daz won't whiten it and Omo won't brighten it, bugger it." Sounds like you've ordered a can of Best Universal Grit, Grime and Effluent Remover.

Edit: Corrected Mr Wedlock's immortal lyric
Last edited by EC8010 on Thu Nov 19, 2026 11:09 am, edited 1 time in total.
Zenith
Posts: 1874
Joined: Sat Oct 22, 2022 9:06 pm

Re: Beware PSU current limits...

Post by Zenith »

Specmaster wrote: ↑Wed Nov 18, 2026 2:46 pm Yes, I will do that. I'm also waiting on some magical solution that also removed all battery corrosion and is supposed to return contacts, etc., to their former glory. Sounds too good to be true, but we shall see.
White vinegar and bicarbonate of soda. According to the Daily Express (aka The Meghan and Harry Times, formerly The Kardashian News) it's a panacea with magical properties and dealing with your problem should be well within its measure. It will have those contacts factory fresh, or not, as the case may be. With flux it won't break a sweat.
tggzzz
Posts: 2841
Joined: Sat Oct 22, 2022 8:17 pm

Re: Beware PSU current limits...

Post by tggzzz »

I thought it was lemon juice. Since it is the Daily sExpress (prop Richard Desmond who also runs Readers' Wives and Asian Babes), I expect the lemon has to be expressed squeezed between the thighs of a voluptuous female.

Is the Express for or against Meghan? Could be either, since they make money from either type of story, and are adept at doing a reverse ferret.
https://en.wikipedia.org/wiki/Reverse_ferret
User avatar
AVGresponding
Posts: 788
Joined: Sat Oct 22, 2022 7:30 pm
Location: The Yorkshire

Re: Beware PSU current limits...

Post by AVGresponding »

There's a decent thread on this on EEVblog as well: https://www.eevblog.com/forum/testgear/ ... teristics/

My Siglent 3303 is pretty good for the most part, but you mustn't connect your load while the output is on, or the output crapacitor will dump its charge into the load. This is under 100µF at a guess, but still plenty for sensitive components.
It's not unique behaviour; my HP and Agilent psus will do the same, though they seem to have a bit less crapacitance.


EDIT: On the removal of schmoo, I use spirit vinegar for alkali contaminants, washed off with 70% IPA afterwards. Haven't had a serious acid contamination yet, but I'd happily use bicarb on that, washed off with 70% IPA again. If you really feel the need, a final deionised water wash; it's cheap enough.
nuqDaq yuch Dapol?

(╯°□°)╯︵ ┻━┻
tggzzz
Posts: 2841
Joined: Sat Oct 22, 2022 8:17 pm

Re: Beware PSU current limits...

Post by tggzzz »

David Hess notes ~50µF per rated amp in decent PSUs. That corresponds with the value I see in my "good" PSUs, with the "bad" and "ugly" being much higher.
User avatar
Specmaster
Posts: 1233
Joined: Sat Oct 22, 2022 8:13 pm
Location: Chelmsford, UK

Re: Beware PSU current limits...

Post by Specmaster »

tggzzz wrote: ↑Wed Nov 18, 2026 11:24 pm I thought it was lemon juice. Since it is the Daily sExpress (prop Richard Desmond who also runs Readers' Wives and Asian Babes), I expect the lemon has to be expressed squeezed between the thighs of a voluptuous female.

Is the Express for or against Meghan? Could be either, since they make money from either type of story, and are adept at doing a reverse ferret.
https://en.wikipedia.org/wiki/Reverse_ferret
I have in the past used both lemon juice and white vinegar but without much success. Maybe it depends on how long the leakage has been present for?

Bearing in mind that the calculators I'm working on are from the 70s and the batteries are internal and the owner's manuals make reference to the batteries beyond the need for them to charged for about 2 to 3 hours, it's quite possible that the uninitiated users just left them tucked away in a drawer once the batteries no longer held charge and they considered plugging the charger in a bit of faff as the LCD models began to gain popularity in the 80s. That means they could have been sitting around quietly somewhere for about 40+ years.
Who let Murphy in?

Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Advance-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi-Heathkit-Duratool
Zenith
Posts: 1874
Joined: Sat Oct 22, 2022 9:06 pm

Re: Beware PSU current limits...

Post by Zenith »

tggzzz wrote: ↑Wed Nov 18, 2026 11:24 pm I thought it was lemon juice. Since it is the Daily sExpress (prop Richard Desmond who also runs Readers' Wives and Asian Babes), I expect the lemon has to be expressed squeezed between the thighs of a voluptuous female.

Is the Express for or against Meghan? Could be either, since they make money from either type of story, and are adept at doing a reverse ferret.
https://en.wikipedia.org/wiki/Reverse_ferret
Dunno. I skim through the on line version. It's a tat rag. it beats me why people would be interested in the small doings of people who are famous for being famous, (celebrities), but they must be or they they wouldn't publish so much of it. The tone is usually sensational. e.g. if a bit of windy weather is forecast it's a hurricane which will lay low the mountains of the earth, and light snow is a "650 mile long snow bomb".
tggzzz
Posts: 2841
Joined: Sat Oct 22, 2022 8:17 pm

Re: Beware PSU current limits...

Post by tggzzz »

Yes on all counts.

Retch (oops Reach) newspapers seem to use WXCharts as their source of weather information. They appear to specialise in "plausible worst case" forecasts, to the extent that I suspect that's their main source of income.
User avatar
EC8010
Posts: 628
Joined: Mon Sep 08, 2025 12:41 am

Re: Beware PSU current limits...

Post by EC8010 »

Ever since Michael Fish got it a bit wrong, weather forecasters have exaggerated possible weather.

Fully agreed on abhorrence of slebs. In broadcast (both Beeb and ITV), there were people who thought they were slebs. Mostly, their only skill was being decorative and able to read Autocue. They were obnoxious.
tggzzz
Posts: 2841
Joined: Sat Oct 22, 2022 8:17 pm

Re: Beware PSU current limits...

Post by tggzzz »

EC8010 wrote: ↑Thu Nov 19, 2026 11:16 am Ever since Michael Fish got it a bit wrong, weather forecasters have exaggerated possible weather.
Decent ones haven't been financially incentivised to increase page impressions. WXCharts/Retch/et al use any and all disreputable techniques to increase impressions.

Having noted that, according to Private Eye[1], Retch and similar have done a reverse ferret in that they finally realise that real local journalism might be what people are prepared to look at.

[1] the best UK source of what people/companies/organisations/governments would like to keep out of sight.
User avatar
EC8010
Posts: 628
Joined: Mon Sep 08, 2025 12:41 am

Re: Beware PSU current limits...

Post by EC8010 »

When I was a kid, my dad used to buy Private Eye. I think it was a bit different then to how it is now. There used to be a record stapled to the Christmas issue whose humour would definitely be beyond the pale now. I remember there was a cartoon strip about mythical Australian hero Barry McKenzie, written by Dame Edna Everage. I do occasionally buy Private Eye now.
tggzzz
Posts: 2841
Joined: Sat Oct 22, 2022 8:17 pm

Re: Beware PSU current limits...

Post by tggzzz »

Jack Hargreaves programmes are now prefaced with warnings about smoking.

Kids cannot understand that Til Death Us Do part was designed to ridicule Alf Garnett, not to admire him.

Having said that, some forms of humour deserve to be extinct. Bernard Manning is a poster child.
tautech
Posts: 252
Joined: Sun Oct 23, 2022 8:24 am
Location: Auckland NZ
Contact:

Re: Beware PSU current limits...

Post by tautech »

AVGresponding wrote: ↑Thu Nov 19, 2026 9:03 am There's a decent thread on this on EEVblog as well: https://www.eevblog.com/forum/testgear/ ... teristics/

My Siglent 3303 is pretty good for the most part, but you mustn't connect your load while the output is on, or the output crapacitor will dump its charge into the load. This is under 100µF at a guess, but still plenty for sensitive components.
It's not unique behaviour; my HP and Agilent psus will do the same, though they seem to have a bit less crapacitance.
The SPD3303X & X-E have 2x 470uF paralleled across outputs 1&2.
Dunno about the earlier C models.....
Siglent Distributor NZ, TE Enabler
User avatar
Specmaster
Posts: 1233
Joined: Sat Oct 22, 2022 8:13 pm
Location: Chelmsford, UK

Re: Beware PSU current limits...

Post by Specmaster »

EC8010 wrote: ↑Thu Nov 19, 2026 11:16 am Ever since Michael Fish got it a bit wrong, weather forecasters have exaggerated possible weather.
In other words, they are covering their arses, in case it is really bad. Then if people ignore their forecasts, they can turn round and say, 'You were warned.'
Who let Murphy in?

Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Advance-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi-Heathkit-Duratool
Zenith
Posts: 1874
Joined: Sat Oct 22, 2022 9:06 pm

Re: Beware PSU current limits...

Post by Zenith »

tggzzz wrote: ↑Thu Nov 19, 2026 6:37 pm Jack Hargreaves programmes are now prefaced with warnings about smoking.
The picture of I. K. Brunel posed with in front of the launching chains of the Great Eastern with stove pipe hat and cigar in mouth, (he chain smoked cigars), has been edited to show him without the cigar. This is just crap. He was a great engineer, he was obviously addicted to nicotine, he was a man of his time, and chain smoking cigars almost certainly contributed to his premature demise. I can see that smoking cigars is to be discouraged, and chain smoking cigars is not going to make you a great engineer, but credit people with an atom of sense and don't edit history.

Jack Hargreaves was a shocking example of a human being, thoroughly atavistic and he appealed to a primitive hunter gatherer instinct we should try to repress, He appeared on film catching fish and killing small furry creatures and our feathered friends with a shotgun. There was something matter of fact and even triumphalist about it.

Going for walk round with him or an afternoon's fishing, would have been a memorable day.
tggzzz wrote: ↑Thu Nov 19, 2026 6:37 pm Kids cannot understand that Til Death Us Do part was designed to ridicule Alf Garnett, not to admire him.
It worked on at least two levels. Alf coming out with things that the bien pensant thought impossibly awful, and many thinking it was about the way they thought, but crudely expressed.
tggzzz wrote: ↑Thu Nov 19, 2026 6:37 pm Having said that, some forms of humour deserve to be extinct. Bernard Manning is a poster child.
Not at all. What's humour worth if it doesn't explore contradictions, nonsenses and assumptions which can be challenged and that can be offensive? I recall he had a clash with what I thought was the completely humourless Joan Rivers.

He was a thing of his time. He came to fame in the 70s along with several others including Charlie Williams (a black Yorkshireman) and Frank Carson. They were working men's clubs acts that found their way onto TV and a national audience. One of the saddest and most poignant things I came across on TV was Frank Carson talking about the troubles in Northern Ireland.
User avatar
EC8010
Posts: 628
Joined: Mon Sep 08, 2025 12:41 am

Re: Beware PSU current limits...

Post by EC8010 »

tggzzz wrote: ↑Thu Nov 19, 2026 6:37 pm Kids cannot understand that Til Death Us Do part was designed to ridicule Alf Garnett, not to admire him.
I did not understand that at the time, and I'm still not convinced by that justification. I thought it was an unpleasant programme featuring an unpleasant man.

I've no idea who Jack Hargreaves is/was, but tweaking history (Brunel's cigar) to make it acceptable to modern tastes is blasphemy! How can anyone learn from the past if its horrors are concealed?
tggzzz
Posts: 2841
Joined: Sat Oct 22, 2022 8:17 pm

Re: Beware PSU current limits...

Post by tggzzz »

EC8010 wrote: ↑Thu Nov 19, 2026 10:24 pm
tggzzz wrote: ↑Thu Nov 19, 2026 6:37 pm Kids cannot understand that Til Death Us Do part was designed to ridicule Alf Garnett, not to admire him.
I did not understand that at the time, and I'm still not convinced by that justification. I thought it was an unpleasant programme featuring an unpleasant man.
When, as a kid, I watched it in the 60s, Alf Garnett came across as a thoroughly boorish but easily recognisable man. He was always out-argued and out-witted by the youngsters, and mostly ignored by his wife. All that enraged him and made his pronouncements even more ridiculous.

In the early 70s I remember Johnny Speight discussing the character he created, clearly stating Alf was a figure of ridicule. Clearly that wasn't a recent retrofit rationalisation!
I've no idea who Jack Hargreaves is/was, but tweaking history (Brunel's cigar) to make it acceptable to modern tastes is blasphemy! How can anyone learn from the past if its horrors are concealed?
I first came across Hargreaves as the grandfather figure in the children's programme How!, which told kids how things they could see in the world worked and were constructed. (The other characters were mother, father, elder brother). I remember his segments being interesting and aimed at demystifying the world. What I didn't know was that he invented the show, and was even involved in setting up commercial TV in the 50s.

He later did programmes about how the countryside worked, describing himself as "just a chap who started talking about the Countryside by accident and found that I as when I thought I was going to do six programs and I did it for 28 years but I never once crusaded for anything and I just want everybody to understand it that's all it's not for me to settle anything I can't it's not for me even to express an opinion I never expressed an opinion on the program I never said to anybody you should do so and so I always said this is what we did and this is what happened"

Such programmes are on CcRewind channel 81 daily (Old Country), and TVTP channel 82 weekly (Out of Town). They can be a bit repetitive, but there is usually an unexpected revealing snippet in a programme. Typically he shows how animals are trained and utilised, how he makes pony carts, household tools, used (as a kid) to catch his lunch (rabbits, pigeons, fish) and why that was so important before farms were mechanised.

Looking back now, it strikes me that How and Out of Town stimulated me into doing and creating things, both professionally and as hobbies.

https://www.youtube.com/watch?v=zS1-aoUNBbE 3:14 long
tautech
Posts: 252
Joined: Sun Oct 23, 2022 8:24 am
Location: Auckland NZ
Contact:

Re: Beware PSU current limits...

Post by tautech »

Zenith wrote: ↑Thu Nov 19, 2026 9:51 pm ................

Jack Hargreaves was a shocking example of a human being, thoroughly atavistic and he appealed to a primitive hunter gatherer instinct we should try to repress,
Why just why ?
Some with skill can provide food for their family this way.
Sure in some parts of the world hunting is only for the elite but not all.
Zenith wrote: ↑Thu Nov 19, 2026 9:51 pm He appeared on film catching fish and killing small furry creatures and our feathered friends with a shotgun. There was something matter of fact and even triumphalist about it.

Going for walk round with him or an afternoon's fishing, would have been a memorable day.
Nothing special about this as all of us in NZ can participate in fresh and saltwater fishing and hunt a myriad of game species with rifle or shotgun or even be part of pest control programmes.

Shooters rights are a thing here but shouldn't be, as in a civilised society a firearm is just another tool like a hammer, screwdriver or knife, all capable of mortal injury just as a vehicle is.
Siglent Distributor NZ, TE Enabler
User avatar
AVGresponding
Posts: 788
Joined: Sat Oct 22, 2022 7:30 pm
Location: The Yorkshire

Re: Beware PSU current limits...

Post by AVGresponding »

tautech wrote: ↑Thu Nov 19, 2026 7:20 pm
AVGresponding wrote: ↑Thu Nov 19, 2026 9:03 am There's a decent thread on this on EEVblog as well: https://www.eevblog.com/forum/testgear/ ... teristics/

My Siglent 3303 is pretty good for the most part, but you mustn't connect your load while the output is on, or the output crapacitor will dump its charge into the load. This is under 100µF at a guess, but still plenty for sensitive components.
It's not unique behaviour; my HP and Agilent psus will do the same, though they seem to have a bit less crapacitance.
The SPD3303X & X-E have 2x 470uF paralleled across outputs 1&2.
Dunno about the earlier C models.....
Yikes, that's a bit more than I expected
nuqDaq yuch Dapol?

(╯°□°)╯︵ ┻━┻
User avatar
AVGresponding
Posts: 788
Joined: Sat Oct 22, 2022 7:30 pm
Location: The Yorkshire

Re: Beware PSU current limits...

Post by AVGresponding »

Specmaster wrote: ↑Thu Nov 19, 2026 9:24 pm
EC8010 wrote: ↑Thu Nov 19, 2026 11:16 am Ever since Michael Fish got it a bit wrong, weather forecasters have exaggerated possible weather.
In other words, they are covering their arses, in case it is really bad. Then if people ignore their forecasts, they can turn round and say, 'You were warned.'
We even get warnings through our Northern Shithole Council email
nuqDaq yuch Dapol?

(╯°□°)╯︵ ┻━┻
Post Reply