Tek 475 chop problem
Posted: Tue Feb 14, 2023 7:59 pm
I have a problem with chop on a Tek 475.
The backstory is that I've had a Tek 475 for about 14 years and used it a fair bit over that time. It's played up twice and been fixed, but it occurred it would be good to have a parts mule.
Last summer I bought a parts mule, but it was so good it seemed a shame to use it for that. It worked with a few problems, such as the timebase vernier knob had had a blow and the shaft was bent, and the A trigger pot had also had a blow.
A couple of months later I came across a proper parts mule for a parts mule price. No display, but it was complete. I decided to fix the second Tek 475.
I sorted out the timebase vernier problem and changed the trigger pot. I checked the supply lines and set the 50V line to 50V within a few mV. All the supply lines were within the spec in the manual. A heatsink on a preamp hybrid was missing and was replaced with one from the parts mule.
There were problems with the vertical sensitivity and AC/GND/DC switch, sorted out by the careful application of Deoxit. Some of the pots weren't smooth, but that went away with use.
Not much left to do but check and adjust the calibration.
Then I noticed a peculiar problem. On channel 1, channel 2 and alt, everything works as expected. On chop there's interference between the two channels.
Set up a square wave trace from the calibrator on one channel and set the other to DC and on alt there are two independent traces. On chop there's also a smaller square wave on what should be a straight line. Set both channels to GND and on alt each can be moved without affecting the other. On chop, move one channel up a few divisions, and the other moves down a good part of a division.
There's no sign of anything like that with the good scope.
I investigated the channel switching circuitry.
The two channels are switched by the hybrid U370.
Pin 1 hi, Pin 12 lo, Pin13 hi, Pin16 lo = Channel 1 selected.
Pin 1 lo, Pin 12 hi, Pin13 lo, Pin16 hi = Channel 2 selected.
Pin 1 hi, Pin 12 hi, Pin13 lo, Pin16 lo = Add channels 1 & 2.
On Alt, alternate sweeps of the timebase select ch1 and ch2 alternately. On Chop the scope switches between channels at about 1MHz.
The waveforms of the defective scope are the same as on the good scope, but the chop frequency is a little lower at 1.8MHz. Voltages and resistor values have been checked and look spot on.
Changing the channel switching hybrid makes no difference and the one from the bad scope works in the good scope.
For those following along here's a jpg of the vertical channel switching circuit.
Removing the 7400 in the chop oscillator and injecting signals from a function generator does not fix the problem. Signals were injected at pin 11 and pin 6 of the 7400. Altering the phase difference between the two signals made the problem slightly better or slightly worse. Altering the frequency between 1.6MHz and 4MHz makes no difference, which suggests the frequency is not critical.
The chop blank waveform looks the same between the good and bad scope.
I'm wondering if this might be a problem with the z axis modulation circuitry.
Has anyone seen anything like this before?
The backstory is that I've had a Tek 475 for about 14 years and used it a fair bit over that time. It's played up twice and been fixed, but it occurred it would be good to have a parts mule.
Last summer I bought a parts mule, but it was so good it seemed a shame to use it for that. It worked with a few problems, such as the timebase vernier knob had had a blow and the shaft was bent, and the A trigger pot had also had a blow.
A couple of months later I came across a proper parts mule for a parts mule price. No display, but it was complete. I decided to fix the second Tek 475.
I sorted out the timebase vernier problem and changed the trigger pot. I checked the supply lines and set the 50V line to 50V within a few mV. All the supply lines were within the spec in the manual. A heatsink on a preamp hybrid was missing and was replaced with one from the parts mule.
There were problems with the vertical sensitivity and AC/GND/DC switch, sorted out by the careful application of Deoxit. Some of the pots weren't smooth, but that went away with use.
Not much left to do but check and adjust the calibration.
Then I noticed a peculiar problem. On channel 1, channel 2 and alt, everything works as expected. On chop there's interference between the two channels.
Set up a square wave trace from the calibrator on one channel and set the other to DC and on alt there are two independent traces. On chop there's also a smaller square wave on what should be a straight line. Set both channels to GND and on alt each can be moved without affecting the other. On chop, move one channel up a few divisions, and the other moves down a good part of a division.
There's no sign of anything like that with the good scope.
I investigated the channel switching circuitry.
The two channels are switched by the hybrid U370.
Pin 1 hi, Pin 12 lo, Pin13 hi, Pin16 lo = Channel 1 selected.
Pin 1 lo, Pin 12 hi, Pin13 lo, Pin16 hi = Channel 2 selected.
Pin 1 hi, Pin 12 hi, Pin13 lo, Pin16 lo = Add channels 1 & 2.
On Alt, alternate sweeps of the timebase select ch1 and ch2 alternately. On Chop the scope switches between channels at about 1MHz.
The waveforms of the defective scope are the same as on the good scope, but the chop frequency is a little lower at 1.8MHz. Voltages and resistor values have been checked and look spot on.
Changing the channel switching hybrid makes no difference and the one from the bad scope works in the good scope.
For those following along here's a jpg of the vertical channel switching circuit.
Removing the 7400 in the chop oscillator and injecting signals from a function generator does not fix the problem. Signals were injected at pin 11 and pin 6 of the 7400. Altering the phase difference between the two signals made the problem slightly better or slightly worse. Altering the frequency between 1.6MHz and 4MHz makes no difference, which suggests the frequency is not critical.
The chop blank waveform looks the same between the good and bad scope.
I'm wondering if this might be a problem with the z axis modulation circuitry.
Has anyone seen anything like this before?