Hitachi V-525 oscilloscope
Posted: Sun Oct 30, 2022 4:01 pm
This came from a rally. It had a sticker on it saying £10. At first I thought it was an Hitachi 20MHz scope. It looked clean, with no missing knobs or signs of being knocked about, and I haven't had a Hitachi scope before. Then I spotted it was 50MHz, so I parted with the tenner without haggling. It turns out they were sold by RS at one time and it has an RS label on the back - not rebadging but one of their preferred standard products.
When I got it home I powered it up and it produced a display with cursors. I fed each channel a square wave. Channel 2 was what was expected, but Channel 1 had what looked like a horrible compensation problem.
I found an operation manual with cct diagrams on the web, but no proper service manual. There is a service manual for the V425 and the circuits are quite similar. I checked the supply line voltages 75V, 8V, 5V and -8V for level and ripple. All pretty much spot on and with about 15mV PTP noise and ripple on the low voltage lines. I was wondering what could cause that sort of overshoot, maybe a dry joint.
The square wave from the calibrator was fed into each channel via a 10X probe and it was checked with a scope at various points through the two channels. Channel 2 produced a square wave as expected. Channel produced a square wave at the BNC connector and at the connector P2, but at the junction of R15 and C16 and thereafter it was distorted. Switching the attenuator stages made no difference to the shape of the waveform.
I wondered what the LF411 was doing. Then it dawned that at DC and low frequency C16 would pass nothing and the LF411 took care of things.
I tried with a DC signal on both channels. Channel 2 behaved normally but Channel 1 showed a third of the expected reading. I measured the resistance from the BNC input to earth on both channels. Channel 2 was 1 MegOhm, Channel 1 was over 3 MegOhms. The culprit had to be R2, a 47R 1206 SMD resistor on the back of the board (it's a single-sided board).
I removed the Y-amp board, and sure enough R2 was blackened. There's a plastic sheet to insulate the bottom of the circuit board from a metal plate forming part of the body of the scope. There was a small brown mark on the plastic corresponding to the position of R2. It looks as if it had overheated.
R2 was replaced and the Y-amp board was checked with a square wave and scope, outside the V-525. There was no need to have the board powered up for that check. Each channel showed a square wave at C16/C116. I cleaned up the timebase rotary switch with IPA and then contact cleaner, because it was fairly easy. It seemed pointless to clean up other switches and controls. It would have involved a lot of dismantling and they were showing no problems.
The scope was reassembled and the problem had gone. It was set up adapting the calibration instructions for the V-425. I thought of re-capping it, or replacing just the main smoothing caps, but decided not to. There were no problems with ripple and I have the impression the scope hasn't had much use.
R2 must have been overheated, but I really can't understand how. Nothing else in that area showed any sign of damage or drifted values. It's hard to see how it could be made to pass enough current to overheat. Possibly with a high voltage applied to the BNC connector, one end could have shorted through the plastic sheet, but the plastic sheet seems sound and there are taller points on the bottom of the board.
Thoughts...
A very nice analogue scope which seems to be a genuine 50MHz. Good build quality and it uses a minimum of special parts. Good quality Alps switches. It has a very sharp display at all points on the screen, which is unusual. Removing the board involved disconnecting plugs but no de-soldering.
The knobs have internal splines which engage with splines on the control shafts. They are a push fit. Some took more force to remove than I was comfortable applying.
I don't much like the cursor display. It's a thing I find only occasionally useful and there's no switch to turn it off. It can be dimmed away using a screwdriver to turn a front panel brightness preset pot. It also disappears when using the strange Ch1 Alt Mag feature.
Ch1 Alt Mag produces the Ch1 display normally and with a 10X expanded time base three divisions below. The two displays cannot be shifted independently.
All in all a nice 50MHz analogue scope and I don't begrudge the tenner.
Here it is pictured in the classic ebay "working" pose. Note that other items of test equipment, cables, leads, probes, manuals, jigs, tools, manuals, blueprints, spares etc are NOT included.
When I got it home I powered it up and it produced a display with cursors. I fed each channel a square wave. Channel 2 was what was expected, but Channel 1 had what looked like a horrible compensation problem.
I found an operation manual with cct diagrams on the web, but no proper service manual. There is a service manual for the V425 and the circuits are quite similar. I checked the supply line voltages 75V, 8V, 5V and -8V for level and ripple. All pretty much spot on and with about 15mV PTP noise and ripple on the low voltage lines. I was wondering what could cause that sort of overshoot, maybe a dry joint.
The square wave from the calibrator was fed into each channel via a 10X probe and it was checked with a scope at various points through the two channels. Channel 2 produced a square wave as expected. Channel produced a square wave at the BNC connector and at the connector P2, but at the junction of R15 and C16 and thereafter it was distorted. Switching the attenuator stages made no difference to the shape of the waveform.
I wondered what the LF411 was doing. Then it dawned that at DC and low frequency C16 would pass nothing and the LF411 took care of things.
I tried with a DC signal on both channels. Channel 2 behaved normally but Channel 1 showed a third of the expected reading. I measured the resistance from the BNC input to earth on both channels. Channel 2 was 1 MegOhm, Channel 1 was over 3 MegOhms. The culprit had to be R2, a 47R 1206 SMD resistor on the back of the board (it's a single-sided board).
I removed the Y-amp board, and sure enough R2 was blackened. There's a plastic sheet to insulate the bottom of the circuit board from a metal plate forming part of the body of the scope. There was a small brown mark on the plastic corresponding to the position of R2. It looks as if it had overheated.
R2 was replaced and the Y-amp board was checked with a square wave and scope, outside the V-525. There was no need to have the board powered up for that check. Each channel showed a square wave at C16/C116. I cleaned up the timebase rotary switch with IPA and then contact cleaner, because it was fairly easy. It seemed pointless to clean up other switches and controls. It would have involved a lot of dismantling and they were showing no problems.
The scope was reassembled and the problem had gone. It was set up adapting the calibration instructions for the V-425. I thought of re-capping it, or replacing just the main smoothing caps, but decided not to. There were no problems with ripple and I have the impression the scope hasn't had much use.
R2 must have been overheated, but I really can't understand how. Nothing else in that area showed any sign of damage or drifted values. It's hard to see how it could be made to pass enough current to overheat. Possibly with a high voltage applied to the BNC connector, one end could have shorted through the plastic sheet, but the plastic sheet seems sound and there are taller points on the bottom of the board.
Thoughts...
A very nice analogue scope which seems to be a genuine 50MHz. Good build quality and it uses a minimum of special parts. Good quality Alps switches. It has a very sharp display at all points on the screen, which is unusual. Removing the board involved disconnecting plugs but no de-soldering.
The knobs have internal splines which engage with splines on the control shafts. They are a push fit. Some took more force to remove than I was comfortable applying.
I don't much like the cursor display. It's a thing I find only occasionally useful and there's no switch to turn it off. It can be dimmed away using a screwdriver to turn a front panel brightness preset pot. It also disappears when using the strange Ch1 Alt Mag feature.
Ch1 Alt Mag produces the Ch1 display normally and with a 10X expanded time base three divisions below. The two displays cannot be shifted independently.
All in all a nice 50MHz analogue scope and I don't begrudge the tenner.
Here it is pictured in the classic ebay "working" pose. Note that other items of test equipment, cables, leads, probes, manuals, jigs, tools, manuals, blueprints, spares etc are NOT included.