Solartron 7150 Plus
Posted: Tue Aug 20, 2024 4:32 pm
I already have one, which is currently on the operating table and under the knife. It does strange things when powered up, sometimes it powers up properly and sometimes it just produces strange patterns on the LCD display. What's shown on the LCD changes when the unit is tilted. I suspect it's probably just a bad connection. It also has the two dim yellow LEDs which make the display hard to read and which need to be replaced with white LEDs. The complicated and expensive IEC connector with filter, fuse and mains voltage selector has failed and needs replacing. It is a work in progress.
I bought a second one from synx508 at the Rugby swapmeet last Sunday. He assured me it works and he has an honest face.
I tried it and it does work.
I have a fairly new Brymen 867s and a cheap voltage standard based on a Maxim MX584KH, which are in agreement within their specified limits. The Solartron's readings came out about mid way between the two.
I have a few Weston Standard Cells I bought a couple of years back for £3 for the lot. I wasn't going to pay any more than a very small amount, because they are very easily damaged. Also they are made of highly toxic materials, cadmium and mercury, and you take on something of a responsibility to dispose of them as hazardous waste. They really aren't supposed to supply any current at all. Connecting them to a 20k/V meter will certainly degrade them. It's easy to imagine someone doing that somewhere along the way. I did check them briefly with a 10MOhm input resistance DMM. I gather that this is just about acceptable, and there's a procedure for measuring their internal resistance along those lines.
Anyway, the Solartron 7150 Plus has a 10 gigaOhm input resistance on its lowest ranges, which should have the standard cell providing 100 pA, and that's generally considered to be fully safe. So I tried it against the standard cells. A couple of the units have two standard cells. I think they for were some balanced arrangement.
The results in Volts are as follows, from left to right..
1.018572 &
1.018579
1.018493 &
1.018510
1.018461
1.018540
They have thermometers and the cases are supposed to be filled with an inert mineral oil, as a thermal buffer, to prevent rapid changes in temperature. They are temperature sensitive and there's a formula to correct from the voltage certified at 20C to the voltage expected at the current temperature. They were supposed to be left to settle for a time before use, if they were moved.
There's a fifth unit shown in a box with two standard resistances, which came in the same lot. That reads 0.915877 V. It's useless. When they were in use as a practical standard, they tended to drop very slightly in voltage with age and had to be sent off to a lab for calibration. The lab would have extra carefully managed and compared standard cells. If the voltage of the cells they were sent had dropped below 1.0183 V the lab wouldn't certify them, as they were considered too unstable.
An enormous amount of R&D went into Weston cells. They had to be treated very carefully and they must have cost a fortune to buy and maintain. It's a fascinating piece of technological history.
It was interesting playing safely with the standard cells. I suppose they'd be a good enough voltage reference which could be taken as 1.0185 Volts, although not particularly practical. I don't know where the Solartron fits in with this, as it could have drifted away from its calibration slightly. I suppose it's the most accurate piece of equipment, and the standard cells have drifted down very slightly. I believe when they were freshly manufactured, they had a voltage which varied slightly around 1.01865 Volts. There were two sorts of standard cells, saturated and unsaturated. It's interesting that the cells in the double units are quite close, probably because they were kept in identical conditions throughout their lives. There are probably all sorts of things you have to do when measuring voltages to a resolution of a few µV to avoid spurious results.
I don't think volt nuttery is a good thing to be drawn into and I've ventured onto a slippery slope.
I bought a second one from synx508 at the Rugby swapmeet last Sunday. He assured me it works and he has an honest face.
I tried it and it does work.
I have a fairly new Brymen 867s and a cheap voltage standard based on a Maxim MX584KH, which are in agreement within their specified limits. The Solartron's readings came out about mid way between the two.
I have a few Weston Standard Cells I bought a couple of years back for £3 for the lot. I wasn't going to pay any more than a very small amount, because they are very easily damaged. Also they are made of highly toxic materials, cadmium and mercury, and you take on something of a responsibility to dispose of them as hazardous waste. They really aren't supposed to supply any current at all. Connecting them to a 20k/V meter will certainly degrade them. It's easy to imagine someone doing that somewhere along the way. I did check them briefly with a 10MOhm input resistance DMM. I gather that this is just about acceptable, and there's a procedure for measuring their internal resistance along those lines.
Anyway, the Solartron 7150 Plus has a 10 gigaOhm input resistance on its lowest ranges, which should have the standard cell providing 100 pA, and that's generally considered to be fully safe. So I tried it against the standard cells. A couple of the units have two standard cells. I think they for were some balanced arrangement.
The results in Volts are as follows, from left to right..
1.018572 &
1.018579
1.018493 &
1.018510
1.018461
1.018540
They have thermometers and the cases are supposed to be filled with an inert mineral oil, as a thermal buffer, to prevent rapid changes in temperature. They are temperature sensitive and there's a formula to correct from the voltage certified at 20C to the voltage expected at the current temperature. They were supposed to be left to settle for a time before use, if they were moved.
There's a fifth unit shown in a box with two standard resistances, which came in the same lot. That reads 0.915877 V. It's useless. When they were in use as a practical standard, they tended to drop very slightly in voltage with age and had to be sent off to a lab for calibration. The lab would have extra carefully managed and compared standard cells. If the voltage of the cells they were sent had dropped below 1.0183 V the lab wouldn't certify them, as they were considered too unstable.
An enormous amount of R&D went into Weston cells. They had to be treated very carefully and they must have cost a fortune to buy and maintain. It's a fascinating piece of technological history.
It was interesting playing safely with the standard cells. I suppose they'd be a good enough voltage reference which could be taken as 1.0185 Volts, although not particularly practical. I don't know where the Solartron fits in with this, as it could have drifted away from its calibration slightly. I suppose it's the most accurate piece of equipment, and the standard cells have drifted down very slightly. I believe when they were freshly manufactured, they had a voltage which varied slightly around 1.01865 Volts. There were two sorts of standard cells, saturated and unsaturated. It's interesting that the cells in the double units are quite close, probably because they were kept in identical conditions throughout their lives. There are probably all sorts of things you have to do when measuring voltages to a resolution of a few µV to avoid spurious results.
I don't think volt nuttery is a good thing to be drawn into and I've ventured onto a slippery slope.