Maplin XG94C RF Signal Generator
Posted: Tue Apr 28, 2026 7:32 pm
These are generic RF signal generators I've seen around since the 90s. This one is branded Maplin, but they were also branded Rapid, Tenma, Altai, Leader and others. There have also been very similar published amateur designs. I'd always wondered about them and noticed they were surprisingly expensive. People seem to like small, neat, modern boxes. I saw it for a tenner at a swapmeet and bought it, mainly with the intention of flipping it.
It's a Colpitts oscillator with six ranges covering 100kHz to 150MHz. The highest range is also labeled as Harmonic 96-450, so it's claimed to be usable on that range using the third harmonic.
I thought it might be interesting to play with and I ought to make sure it worked before selling it.
The frequency accuracy is given as +/- 3% and the output 100mV (approx) up to 35MHz. It doesn't quote an output impedance, but I'd estimate it varies over a range of a few hundred Ohms. I haven't tried measuring it. It isn't a standard 50 Ohm output. It doesn't have a proper attenuator, it has a pot and a switched potential divider giving High and Low ranges.
It has internal AM modulation at about 1kHz and can be externally modulated. There's also a crystal position and sockets for a of crystal up to 15MHz, on the front. The manual discusses how this can be used with an external receiver for accurate "calibration". The inductors and capacitors have no trimmers and the unit is stuck with the calibration it had when it left the factory.
I tried connecting it to a scope and it was more or less accurate. It does have a slow motion drive but the cursor is quite coarse. There are markers on the scale for 455kHz, 4.5MHz and 10.7MHz, all common IF frequencies. The output amplitude is far from constant and is very much lower on the higher ranges. It rises and drops noticeably across the upper scales and harmonics are prominent. I'd guess this is because of stray capacitance and inductance. Probably the most irritating aspect of using it for IF alignment would be that it's hard to set and reset with sufficient accuracy, even using the markers. This can't be done conveniently from the output using a frequency counter. There is a modification which takes the signal from where it enters the output attenuator, through a capacitor, to a BNC socket. So a frequency counter will have an adequate signal. Many frequency counters have an oscilloscope type 1MOhm 20pF input on their low range. The modification is straightforward and so I did it.
With the modification in place it's much easier to set the frequency accurately. For instance it can be set to 455kHz to within about 50Hz without much trouble. It drifts surprisingly little.
Circuit diagram. Sig Gen connected to scope and frequency counter.
Internals of sig gen. Scope screen shot at 455kHz. Scope screen shot at 20MHz.
This is definitely a hobbyist instrument. It lacks internal screening and leakage can easily be picked up on a scanner. The output amplitude varies wildly across the upper ranges. The distortion of the sine wave on the higher ranges is striking. AM modulation is crude. I have an Advance RF sig gen from around 1960 which uses a single ECC81 (12AT7). It is not a lab grade instrument but it has much better screening than the Maplin unit and does have presets to adjust it. There are good reasons why lab standard RF sig gens are expensive.
I believe the Maplin XG94C works as well as it ever did and the modification makes it more useful. It could be used to do IF alignments on radios, which I'd guess was one of its major intended purposes. I must admit I'm not very impressed by it.
It's a Colpitts oscillator with six ranges covering 100kHz to 150MHz. The highest range is also labeled as Harmonic 96-450, so it's claimed to be usable on that range using the third harmonic.
I thought it might be interesting to play with and I ought to make sure it worked before selling it.
The frequency accuracy is given as +/- 3% and the output 100mV (approx) up to 35MHz. It doesn't quote an output impedance, but I'd estimate it varies over a range of a few hundred Ohms. I haven't tried measuring it. It isn't a standard 50 Ohm output. It doesn't have a proper attenuator, it has a pot and a switched potential divider giving High and Low ranges.
It has internal AM modulation at about 1kHz and can be externally modulated. There's also a crystal position and sockets for a of crystal up to 15MHz, on the front. The manual discusses how this can be used with an external receiver for accurate "calibration". The inductors and capacitors have no trimmers and the unit is stuck with the calibration it had when it left the factory.
I tried connecting it to a scope and it was more or less accurate. It does have a slow motion drive but the cursor is quite coarse. There are markers on the scale for 455kHz, 4.5MHz and 10.7MHz, all common IF frequencies. The output amplitude is far from constant and is very much lower on the higher ranges. It rises and drops noticeably across the upper scales and harmonics are prominent. I'd guess this is because of stray capacitance and inductance. Probably the most irritating aspect of using it for IF alignment would be that it's hard to set and reset with sufficient accuracy, even using the markers. This can't be done conveniently from the output using a frequency counter. There is a modification which takes the signal from where it enters the output attenuator, through a capacitor, to a BNC socket. So a frequency counter will have an adequate signal. Many frequency counters have an oscilloscope type 1MOhm 20pF input on their low range. The modification is straightforward and so I did it.
With the modification in place it's much easier to set the frequency accurately. For instance it can be set to 455kHz to within about 50Hz without much trouble. It drifts surprisingly little.
Circuit diagram. Sig Gen connected to scope and frequency counter.
Internals of sig gen. Scope screen shot at 455kHz. Scope screen shot at 20MHz.
This is definitely a hobbyist instrument. It lacks internal screening and leakage can easily be picked up on a scanner. The output amplitude varies wildly across the upper ranges. The distortion of the sine wave on the higher ranges is striking. AM modulation is crude. I have an Advance RF sig gen from around 1960 which uses a single ECC81 (12AT7). It is not a lab grade instrument but it has much better screening than the Maplin unit and does have presets to adjust it. There are good reasons why lab standard RF sig gens are expensive.
I believe the Maplin XG94C works as well as it ever did and the modification makes it more useful. It could be used to do IF alignments on radios, which I'd guess was one of its major intended purposes. I must admit I'm not very impressed by it.