Racal 9916 frequency counter repair
Posted: Tue Dec 06, 2022 11:14 pm
This came from a rally in the summer. It was claimed to be working. It was working on the B range (10Hz-60MHz) but the A range (40MHz -520MHz) was dead.
Opened it up and checked the supply line voltages. The 5V and -5v lines were about as specified in the manual. When the A range is set, part of IC14 is used to divide by two 1MHz derived from the main frequency reference. This 500KHz signal is used to drive Q7 on and off and energise the primary of T1. The voltage in the secondary of T1 is rectified and added to the +5V line to create a 24V supply for IC 26, a wide band hybrid amplifier. It is smoothed by a filter comprising C30 and C31 (4.7uF and 1.5uF tants) and L4, a 3.3uH inductance.
Test point 2 is provided to check the output of this PSU arrangement for IC26, the hybrid amplifier. It was 3.7V with the B range selected and 4.3V with the A range selected, not 24V. Q7 (ZTX 450) was hot to the touch. A scope showed a 500KHz square wave from IC14 driving the base of Q7.
I suspected one of the tants, (C30 and C31) was short circuit, or IC26 an OM335 hybrid circuit had failed, and was drawing heavy current. The replacement parts in that part of the circuit are mostly quite cheap, so I ordered the lot from CPC and had to wait a time for them all to arrive. The OM335 is still available on ebay for about £8, but I could see no point in jumping to conclusions about that.
I desoldered C30 and C31. For some reason it was very hard, even with a desoldering station. I then had the idea of lifting one end of L4 to isolate the OM335. There is no L4, it was replaced by R166 a 68R resistor. This took some time to work out. I had the idea of putting in a dummy load for IC26 which draws 35mA according to its datasheet. 680R was about right. A 3uF 50V electrolytic was put in place across it. This arrangement was connected between the cathodes of D6 and D7 and earth. The instrument was powered on and the voltage across the load was monitored with a scope. 3.7V on range B, 23V and little ripple on range A.
Next I soldered in the replacements for C30 and C31. I soldered the cut lead of R166 back together. I'd rather replace a part than do that, but it was such a pain to remove the two caps that I couldn't face removing and replacing it. It seems to be a sound joint. Tweaking R166 with long nosed pliers couldn't break the joint.
It was still possible that IC26 was faulty.
With everything replaced and checked. the instrument was powered on. 22.9V at TP2 with A range selected, which is within spec, and it read about what was expected when fed from an RF signal generator.
After it had been allowed to warm up, it was set up against a 10MHz GPSDO. It seems to settle to less than 1 part in 10^7 and return there after it's been powered on after being powered off for a time. It has the lesser of the two OCXO options Racal offered. I had thought of replacing the OCXO with something based on one of the Aliexpress cell tower pulls, but if it continues like this I won't.
Piccies
A pdf of the relevant bit of the circuit for anyone inclined to follow. A picture of the circuit board around the 24V PSU area. The two orange blobs in the middle are the new C30 and C31. R166/L4 is above them. The orange/brown long thing above R166 is the OM355. TP2 is between the OM355 and R166. The grey, square thing to the right with an orange dot is T1. Q7 is the closest transistor beneath T1. Diodes D6, D7 and D8 are beneath and to the right of T1. The polarity of a lot of the diodes is indicated incorrectly on the silk screen, which caused some head scratching.
A picture of the 9916 showing as close to 10MHz as I could adjust it.
Final thoughts. A fairly simple repair with no nasty and expensive surprises. The manual is a lot better than nothing, but with a few omissions. Not a thing I found pleasant to work on. This thing is packed full of tants. At least one has died. I suppose most of them could be replaced with electrolytics with a ceramic in parallel, but removing them is a PITA.
Opened it up and checked the supply line voltages. The 5V and -5v lines were about as specified in the manual. When the A range is set, part of IC14 is used to divide by two 1MHz derived from the main frequency reference. This 500KHz signal is used to drive Q7 on and off and energise the primary of T1. The voltage in the secondary of T1 is rectified and added to the +5V line to create a 24V supply for IC 26, a wide band hybrid amplifier. It is smoothed by a filter comprising C30 and C31 (4.7uF and 1.5uF tants) and L4, a 3.3uH inductance.
Test point 2 is provided to check the output of this PSU arrangement for IC26, the hybrid amplifier. It was 3.7V with the B range selected and 4.3V with the A range selected, not 24V. Q7 (ZTX 450) was hot to the touch. A scope showed a 500KHz square wave from IC14 driving the base of Q7.
I suspected one of the tants, (C30 and C31) was short circuit, or IC26 an OM335 hybrid circuit had failed, and was drawing heavy current. The replacement parts in that part of the circuit are mostly quite cheap, so I ordered the lot from CPC and had to wait a time for them all to arrive. The OM335 is still available on ebay for about £8, but I could see no point in jumping to conclusions about that.
I desoldered C30 and C31. For some reason it was very hard, even with a desoldering station. I then had the idea of lifting one end of L4 to isolate the OM335. There is no L4, it was replaced by R166 a 68R resistor. This took some time to work out. I had the idea of putting in a dummy load for IC26 which draws 35mA according to its datasheet. 680R was about right. A 3uF 50V electrolytic was put in place across it. This arrangement was connected between the cathodes of D6 and D7 and earth. The instrument was powered on and the voltage across the load was monitored with a scope. 3.7V on range B, 23V and little ripple on range A.
Next I soldered in the replacements for C30 and C31. I soldered the cut lead of R166 back together. I'd rather replace a part than do that, but it was such a pain to remove the two caps that I couldn't face removing and replacing it. It seems to be a sound joint. Tweaking R166 with long nosed pliers couldn't break the joint.
It was still possible that IC26 was faulty.
With everything replaced and checked. the instrument was powered on. 22.9V at TP2 with A range selected, which is within spec, and it read about what was expected when fed from an RF signal generator.
After it had been allowed to warm up, it was set up against a 10MHz GPSDO. It seems to settle to less than 1 part in 10^7 and return there after it's been powered on after being powered off for a time. It has the lesser of the two OCXO options Racal offered. I had thought of replacing the OCXO with something based on one of the Aliexpress cell tower pulls, but if it continues like this I won't.
Piccies
A pdf of the relevant bit of the circuit for anyone inclined to follow. A picture of the circuit board around the 24V PSU area. The two orange blobs in the middle are the new C30 and C31. R166/L4 is above them. The orange/brown long thing above R166 is the OM355. TP2 is between the OM355 and R166. The grey, square thing to the right with an orange dot is T1. Q7 is the closest transistor beneath T1. Diodes D6, D7 and D8 are beneath and to the right of T1. The polarity of a lot of the diodes is indicated incorrectly on the silk screen, which caused some head scratching.
A picture of the 9916 showing as close to 10MHz as I could adjust it.
Final thoughts. A fairly simple repair with no nasty and expensive surprises. The manual is a lot better than nothing, but with a few omissions. Not a thing I found pleasant to work on. This thing is packed full of tants. At least one has died. I suppose most of them could be replaced with electrolytics with a ceramic in parallel, but removing them is a PITA.