After a while it started showing signs of plausible life, but with completely unsteady readings and no decimal point (shades of Siglent FFTs

So I took it apart and irrigate the many switches with IPA. Much better, decimal points had reappeared, readings were mostly good but there was noise and the 20V scale was offset by up to 0.5V.
Flipped the board and there was clear evidence of simian involvement by the IEC inlet. ...
Couldn't they have soldered the PCB to the connector properly?! Tried my Metcal with a beefy "ground plane" tip - barely touched it! WTF?
Flipped the board again and found why: each connector pin is a solid slab that is self-supporting all the way to the PCB. No shroud, apart from the small external Fluke-blue bit of plastic that slips off?! WTF again.
Look more closely, especially at the other side of the connector (with Fluke blue shroud reinstalled) and see there is a shroud made of - wait for it - perforated card?! Involuntary shudder, as I remember Bulgin connectors.
Remaining problem is the 20V offset. What could be causing that? Remember the concept of select-on-test resistors? Yes, they were a horrible practical necessity at the time. But they were tolerable when soldered in, or if selected by snipped wire links as in the bomb proof Fluke I tore down ant The Other Plaice https://www.eevblog.com/forum/testgear/ ... msg2540730
But, oh no, that would be too difficult to implement. Guess how Fluke did implement it...
... go on, guess ...
... using my favourite technology, a mini solderless breadboard.
Waggle those resistors up and down to dislodge the oxide, and the 20V offset "magically" disappears.
Job done.